2025-01-01 2:20 PM
Hello,
I have a problem with STM32N6. When trying to access any TCM (ITCM/DTCM) the CPU immediately crashes to the Hard_Fault. Do the TCM memories need to be explicitly enabled or something? From my experience from STM32H7, it was never needed. I've looked around RM for N6 but cannot find anything related. I have found in the examples (FLEXMEM_Configurations) that the HardFault shall happen when accesing beyond the configured TCM size, but I'm getting the crash at the base addresses.
uint16_t* _tcmData = (uint16_t*)0x30000000;
int main()
{
... //system init etc..
_tcmData[0] = 0x1234; //Hard_Fault crash here
}
Solved! Go to Solution.
2025-01-04 5:50 AM
The TCM needs to be initialized before use it. Here is a snipped assembler code that you will need to add in the reset handler of your startup_stm32n657xx.s file (replace the Reset_Handler function)
It will allow to correctly initialize at 0 your DTCM - Baseline region at address 0x30000000 and allow LDR operations without PECC error and then hardfault.
Reset_Handler:
ldr r0, =_sstack
msr MSPLIM, r0
ldr r0, =_estack
mov sp, r0 /* set stack pointer */
/* Clear D-TCM */
ldr R0, = 0x30000000
ldr R1, = 0x30020000
mov R2, #0
clear_dtcm:
str R2, [R0]
add R0, R0, #4
cmp R0, R1
bcc clear_dtcm
/* Call the clock system initialization function.*/
bl SystemInit
And with my code:
/* USER CODE BEGIN PV */
uint16_t* _tcmData0 = (uint16_t*)0x30000000;
uint16_t* _tcmData1 = (uint16_t*)0x30000010;
uint16_t* _tcmData2 = (uint16_t*)0x30000020;
/* USER CODE END PV */
...
int main(void)
{
/* USER CODE BEGIN 1 */
/* USER CODE END 1 */
/* MCU Configuration--------------------------------------------------------*/
HAL_Init();
/* USER CODE BEGIN Init */
_tcmData0[0] = 0x1234;
_tcmData1[0] = 0x1234;
_tcmData2[0] = 0x1234;
/* USER CODE END Init */
/* Configure the system clock */
SystemClock_Config();
After that, whatever you do with the TCM memory, you should also define these corresponding regions in the linker file. This is how it should be done.
Remember that on one side there is a STM32N6 device which is a complex SoC. And on the other side there are the compiler tools. For both to match and work, there are these best practices that must be applied.
I hope it will fix your hardfault issues when you use TCM with STM32N6. Let me know.
Best regards,
Romain,
To give better visibility on the answered topics, please click on Accept as Solution on the reply which solved your issue or answered your question.
2025-01-03 9:45 PM
VERY INTERESTING (and confusing)!
I play right now also with the STM32N6 - I tried your issue on on my project... and YES: it can crash in HardFault_Handler (but see below: it depends just on how you compile with debug options!)
First, I thought:
A non-secure vs. secure issue (DTCM 0x30000000 is Secure, 0x20000000 is Non-Secure). So, accessing DTCM as secure when main() is launched as non-secure (potentially) could cause such an issue. But it does NOT!
It just depends on:
How you set Optimization level! VERY STRANGE:
I use an STM32N6 project with FSBL build. It really just depends on how I set the debug level!
Secure and non-secure address location does not have an effect! (it crashes with both address locations if not -Og!)
What is this?
(I have realized already: the STM32N6 is very "delicate": sometimes it works, sometimes not - it seems to depend meanwhile on "how" I generate the code (with vs. without debug optimization, FSBL vs. Appli, ...).
Here the details:
I do also in main():
int main(void)
{
int no = 6;
/* USER CODE BEGIN 1 */
XSPI_RegularCmdTypeDef sCommand = {0};
XSPI_MemoryMappedTypeDef sMemMappedCfg = {0};
uint16_t index = 0;
uint16_t res = 0;
uint32_t address = 0;
__IO uint8_t step = 0;
__IO uint8_t *mem_addr;
#if 1
{
uint16_t* _tcmData = (uint16_t*)0x20000000; //0x30000000 is secure memory, 0x20000000 is non-secure
_tcmData[0] = 0x1234;
if (_tcmData[0] == 0x1234)
testVar = 1;
}
#endifJust toggle debug level for code generation!
This is a debug session with debug level -Og:
This works fine!: I can write and read DTCM as secure and non-secure (0x30000000 vs. 0x20000000).
This is a debug session with none, as -O0:
This CRASHES: both addresses accessed result in a HardFault_Handler.
What is this?
I assume: the debugger has an effect on secure vs. non-secure access permission or system configuration done (by debugger instead of code).
(the system behavior should not not have any effect if generated for debugger and/or using debugger to let execute my code)
I am pretty "shocked" being able to replicate this issue (and I am very confused, losing my confidence in this N6 MCU...)
2025-01-04 12:23 AM - edited 2025-01-04 4:52 AM
Very interesting findings @tjaekel .
I can confirm, if I switch to -O2, then it does not crash into the HardFault when accessing the memory. However, this is far from over.
To address your points:
(I have realized already: the STM32N6 is very "delicate": sometimes it works, sometimes not - it seems to depend meanwhile on "how" I generate the code (with vs. without debug optimization, FSBL vs. Appli, ...).
Same here. The N6 goes wild on me like 50 times a day, when I have to physically reconnect the USB to reset the ST-Link and Power. Reset button does not "fix" things.
(I'm using the N6-Nucleo board)
A non-secure vs. secure issue (DTCM 0x30000000 is Secure, 0x20000000 is Non-Secure). So, accessing DTCM as secure when main() is launched as non-secure (potentially) could cause such an issue. But it does NOT!
I'm no expert in this secure/non-secure thing. Actually, it just bothers me, I don't want it, but it is what it is. But I was thinking, my code and data runs from secure (using the Appli LRUN.ld, which places the code and data into the 0x34000400 AXI SRAM, which is secure. When I was checking addresses of peripherals, they were all in the 0x50000000 range, again, secure. So my thinking was - I'm in secure, I shall use secure.
I hope these are mostly ST-link issue, fixable by FW update of ST-link. It is a new product, it is very complex, so yeah, I'm expecting some newborn pains.
Because - when I move my stack into the DTCM, it works.
What does not work for me is ITCM. I just cannot make it run no matter what. I have my code and approach from STM32H7:
This has always worked for me. But not in N6 case, I just always get a HardFault when calling any func placed at ITCM.
2025-01-04 12:44 AM
Also tried accessing the memory through STM32CubeProgrammer. I can access basically any address:
But, when I try to access 0x00000000 or 0x10000000 or 0x20000000 or 0x30000000 it gives me an error.
So I can't check or confirm, what is in the ITCM (if anything).
Please ST, have a look at this.
2025-01-04 4:57 AM - edited 2025-01-04 5:32 AM
Update: I have also tried to load the ITCM mem "directly", without the intermediate step of loading from ROM/RAM and this works, surprisingly. When debugging, I can see the SP switch to 0x10000000 and executes the code in ITCM normally. Really, really weird and unpredictable behaviour. Also it is useless.
Edit: Even better, when I load the ITCM directly, it looks like the "connection" to the ITCM gets somehow "enabled" and as long as I don't disconnect the power, I can then do the normal approach with loading the contents to the ITCM from RAM/ROM. And it works. But, when I disconnect the power, reconnect and upload exactly the same code again -> HardFault.
2025-01-04 5:50 AM
The TCM needs to be initialized before use it. Here is a snipped assembler code that you will need to add in the reset handler of your startup_stm32n657xx.s file (replace the Reset_Handler function)
It will allow to correctly initialize at 0 your DTCM - Baseline region at address 0x30000000 and allow LDR operations without PECC error and then hardfault.
Reset_Handler:
ldr r0, =_sstack
msr MSPLIM, r0
ldr r0, =_estack
mov sp, r0 /* set stack pointer */
/* Clear D-TCM */
ldr R0, = 0x30000000
ldr R1, = 0x30020000
mov R2, #0
clear_dtcm:
str R2, [R0]
add R0, R0, #4
cmp R0, R1
bcc clear_dtcm
/* Call the clock system initialization function.*/
bl SystemInit
And with my code:
/* USER CODE BEGIN PV */
uint16_t* _tcmData0 = (uint16_t*)0x30000000;
uint16_t* _tcmData1 = (uint16_t*)0x30000010;
uint16_t* _tcmData2 = (uint16_t*)0x30000020;
/* USER CODE END PV */
...
int main(void)
{
/* USER CODE BEGIN 1 */
/* USER CODE END 1 */
/* MCU Configuration--------------------------------------------------------*/
HAL_Init();
/* USER CODE BEGIN Init */
_tcmData0[0] = 0x1234;
_tcmData1[0] = 0x1234;
_tcmData2[0] = 0x1234;
/* USER CODE END Init */
/* Configure the system clock */
SystemClock_Config();
After that, whatever you do with the TCM memory, you should also define these corresponding regions in the linker file. This is how it should be done.
Remember that on one side there is a STM32N6 device which is a complex SoC. And on the other side there are the compiler tools. For both to match and work, there are these best practices that must be applied.
I hope it will fix your hardfault issues when you use TCM with STM32N6. Let me know.
Best regards,
Romain,
To give better visibility on the answered topics, please click on Accept as Solution on the reply which solved your issue or answered your question.
2025-01-04 6:15 AM
@RomainR. Thank you! This works.
I realize, that the M55 core is different from M7 core. But I just never knew that the ITCM/DTCM memories need some kind of "init". I've modified your snippet to also init ITCM memory
/* Clear I-TCM */
ldr R0, = 0x10000000
ldr R1, = 0x10010000
mov R2, #0
clear_itcm:
str R2, [R0]
add R0, R0, #4
cmp R0, R1
bcc clear_itcmAnd now DTCM and ITCM are both available, readable and accessible. Awesome, thanks again.
2025-01-04 10:58 AM
Great!
What is the conclusion? For my impression: TCMs have ECC, always on. On startup - it looks like - there is a mismatch between random RAM content and CRC. So, it needs a "clear" cycle, to initialize TCMs first before first use (to bring RAM content and ECC into "sync").
Anyway: an "interesting feature" (compared to H7 MCUs).
2025-01-04 3:18 PM
I cannot stop thinking about this "issue". I want to understand completely what is going on. What I come across is this:
Here the test code I use (in main(), booting in DEV mode = not reading and not booting this code from external octal SPI flash):
{
int i;
uint32_t* _dtcmData = (uint32_t*)0x30000000; //0x30000000 is secure memory, 0x20000000 is non-secure - both OK
for (i = 0; i < 512; i++)
_dtcmData[i] = 0x1234;
if (_dtcmData[0] == 0x1234)
testVar = 1;
uint32_t* _itcmData = (uint32_t*)0x00000000; //0x00000000 is code non-secure, 0x10000000 is code secure - both OK
for (i = 0; i < 512; i++)
_itcmData[i] = 0x47702001;
////if (_itcmData[0] == 0x1234) //HardFault_Handler - WHY? --> it is an UDF instruction generated when 0x00000000!
if (_itcmData[1] == 0x47702001) //THIS WORKS! compiler generates warning about pointer = 0 and any index in array
testVar = 2;
/* it works with address 0x00000004 - but not with reading from address 0x00000000! - compiler creates UDF instruction! */
uint32_t* xPtr = (uint32_t*)0x00000004; //0x00000000 not possible - it generates UDF instruction!
/* but possible to modify register and read from address 0x0 - with debugger */
if (*xPtr == 0x47702001) //0x4770, 0x2001 is MOV R0, #1, BX LR instruction code - see below calling it
testVar = 3;
{
/* this works: code can be executed at address 0x00000000 - but not generated (compiler) */
FPTR fptr = (FPTR)0x00000001; //it must be thumb code address!
if (fptr() == 1)
{
testVar = 4; //it works!
}
}
}OK, much clearer now for me (and fine, as long as we have the "constraints" in mind).