2018-10-19 5:31 AM
Temperature compensation for this magnetometer is not working as I expect. When heating device from room temperature to 50-60 degrees Celsius in 3 minutes magnetometer values are changing a lot.
I have 100Hz ODR, high-precision mode, temperature sensor and temperature compensation enabled, offset cancelation disabled.
Anyone having an idea what else should I set up? What is the normal (delta C / minute) which sensor can work with while compensating properly?
Solved! Go to Solution.
2018-10-25 1:41 AM
You have to also enable offset cancellation, to achieve specified offset vs temperature.
2018-10-22 5:28 AM
What it means the values are changing a lot?
Can you please share your results and exact sensor configuration so I can check it?
2018-10-22 5:41 AM
@Miroslav BATEK
Here is the log from the device, raw value is plotted in red. So change is around 100, 50 and 15 mGauss for Z, Y and X axes respectively.
I would be happy if all 3 axes will behave as X but in current setup it gives me false positives.
My initialization code is:
mgn_status_t LSM303AGR_M_Init(void)
{
if (LSM303AGR_M_Check_WhoAmI() == MAGNETO_ERROR)
{
return MAGNETO_ERROR;
}
/* Operating mode selection - power down */
if (LSM303AGR_MAG_W_MD(0, LSM303AGR_MAG_MD_IDLE1_MODE) == MEMS_ERROR)
{
return MAGNETO_ERROR;
}
/* Enable BDU */
if (LSM303AGR_MAG_W_BDU(0, LSM303AGR_MAG_BDU_DISABLED) == MEMS_ERROR)
{
return MAGNETO_ERROR;
}
if (LSM303AGR_M_Set_ODR(ODR_LOW) == MAGNETO_ERROR)
{
return MAGNETO_ERROR;
}
if (LSM303AGR_M_Set_FS(FS_LOW) == MAGNETO_ERROR)
{
return MAGNETO_ERROR;
}
/* Disable Self-test */
if (LSM303AGR_MAG_W_ST(0, LSM303AGR_MAG_ST_DISABLED) == MEMS_ERROR)
{
return MAGNETO_ERROR;
}
/* High-resolution mode */
if (LSM303AGR_MAG_W_LP(0, LSM303AGR_MAG_HR_MODE) == MEMS_ERROR)
{
return MAGNETO_ERROR;
}
/* Low-pass filter */
if (LSM303AGR_MAG_W_LPF(0, LSM303AGR_MAG_LPF_ENABLED) == MEMS_ERROR)
{
return MAGNETO_ERROR;
}
/* Enable TEMP sensor */
if (LSM303AGR_MAG_W_TemperatureSensor(0, LSM303AGR_MAG_TEMP_EN_ENABLE) == MEMS_ERROR)
{
return MAGNETO_ERROR;
}
/* Enable TEMP compensation */
if (LSM303AGR_MAG_W_COMP_TEMP_EN(0, LSM303AGR_MAG_COMP_TEMP_EN_ENABLED) == MEMS_ERROR)
{
return MAGNETO_ERROR;
}
/* Offset cancellation */
/*if (LSM303AGR_MAG_W_OFF_CANC(0, LSM303AGR_MAG_OFF_CANC_ENABLED) == MEMS_ERROR)
{
return MAGNETO_ERROR;
}*/
return MAGNETO_OK;
}I am sure that I have COMP_TEMP_EN = 1 and TEMP_EN[1:0] = 1
2018-10-25 1:32 AM
@Miroslav BATEK did you have any chance to look at it? Is it normal behavior of the sensor or not?
2018-10-25 1:41 AM
You have to also enable offset cancellation, to achieve specified offset vs temperature.
2019-12-07 4:18 AM
I want to use LSM303AGR instead of LIS3MDL to detect a vehicle parking status.
If i get the x-y-z component every second and calculate X=X_new - X_Previous, then new vehicle that comes near sensor is neglected after 1 second and i can't detect a vehicle that parked for 1 hour!
Does "offset cancellation" option do sth like this?
In other word, if i enable offset cancellation, Does it effect on total X(ory or z) components or just their thermal drift?
We’re moving the ST Community to a new platform to give you a better and more reliable community experience.