2017-03-28 8:08 AM
Hi,
is it possible to know the minimum delay required to wait between register writing during the self-test?
I write down my code, to make myself more clear
while(1)
{ BYTE xH = 0; BYTE yH = 0; BYTE zH = 0; //HPF disabling writeRegAccel(BYTE_REG2, WRT_CTRL_REG2_WO_MS); //Minimum delay to swotch from a mode to another delayMs(7); //I set the frequency at 5k376 Hz writeRegAccel(LOW_PWR_MODE_5K376HZ_W_XYZ, WRT_CTRL_REG1_WO_MS); delayMs(7); //I read the 3 H axis //First reading with garbage data xxx1: if(((readRegAccel(STATUS_REG)) & 0x08) == 0x08) { readRegAccel(OUT_X_H); readRegAccel(OUT_Y_H); readRegAccel(OUT_Z_H); }else{ goto xxx1; } //Second reading of the 'real' datas xxx2: if(((readRegAccel(STATUS_REG)) & 0x08) == 0x08) { xH = readRegAccel(OUT_X_H); resultAbsX = abs(xH); yH = readRegAccel(OUT_Y_H); resultAbsY = abs(yH); zH = readRegAccel(OUT_Z_H); resultAbsZ = abs(zH); }else{ goto xxx2; } //Self-test function accel_Test(); //Display refresh shwAccelTest(SHW_ACCEL_RFSH); //Disabling the self-test mode writeRegAccel(BYTE_REG4, WRT_CTRL_REG4_WO_MS); //I clear the memory in case there are garbage readings xxx3: if(((readRegAccel(STATUS_REG)) & 0x08) == 0x08) { readRegAccel(OUT_X_H); readRegAccel(OUT_Y_H); readRegAccel(OUT_Z_H); }else{ goto xxx3; } xxx4: if(((readRegAccel(STATUS_REG)) & 0x08) == 0x08) { readRegAccel(OUT_X_H); readRegAccel(OUT_Y_H); readRegAccel(OUT_Z_H); }else{ goto xxx4; } //HPF reactivation writeRegAccel(BYTE_REG2_HP, WRT_CTRL_REG2_WO_MS); //I set the frequency to 50 Hz writeRegAccel(LOW_PWR_MODE_50HZ_W_XYZ, WRT_CTRL_REG1_WO_MS); delayMs(20); }accel_Test FUNCTION
void accel_Test()
{ BYTE xH_Test = 0; BYTE yH_Test = 0; BYTE zH_Test = 0; BYTE delta_X = 0; BYTE delta_Y = 0; BYTE delta_Z = 0; //I activate the self-test writeRegAccel(BYTE_REG4_SELF_TEST, WRT_CTRL_REG4_WO_MS); //Minimum delay to activate the self-test delayMs(7); //First reading (garbage data) xxx3: if(((readRegAccel(STATUS_REG)) & 0x08) == 0x08) { readRegAccel(OUT_X_H); readRegAccel(OUT_Y_H); readRegAccel(OUT_Z_H); }else{ goto xxx3; } //Second reading of the effective data xxx4: if(((readRegAccel(STATUS_REG)) & 0x08) == 0x08) { xH_Test = readRegAccel(OUT_X_H); resultAbsTestX = abs(xH_Test); yH_Test = readRegAccel(OUT_Y_H); resultAbsTestY = abs(yH_Test); zH_Test = readRegAccel(OUT_Z_H); resultAbsTestZ = abs(zH_Test); }else{ goto xxx4; }//I make a delta between the two reading
delta_X = resultAbsTestX - resultAbsX; delta_Y = resultAbsTestY - resultAbsY; delta_Z = resultAbsTestZ - resultAbsZ; //Delta's absolute of the datas delta_X = abs(delta_X); delta_Y = abs(delta_Y); delta_Z = abs(delta_Z); //Results control if(delta_X > KMIN_DELTA_TEST_ACC || delta_Y > KMIN_DELTA_TEST_ACC || delta_Z > KMIN_DELTA_TEST_ACC ) { //Result OK }else{ //Result error } }If I don't put those delay of 7ms, the accelerometer will give me back wrong datas, like 0xff or datas from before the self-test was deactivated.
Thus, is there a minimum delay I need to set after I change my frequency or set other registers?
I translated my comments so that the reading will be simplier.
Thank you
#lis2dh12 #self-test #delay #register2017-03-29 1:23 AM
Recommended delay after the self-test is enabled is 200ms.
You should also discard 2 or 8 samples according to selected mode (see note 7 below)
2017-03-30 2:24 AM
And when I change the HPF mode and frequency? Do I need to put a minimum delay here too? Because I noticed that if I don't delay of a minimum of 7 ms when I change those, I get strange readings.
2018-01-10 3:03 PM
ST rep sent a confidential doc specific to the selftest of the LIS2DH that calls for a minimum 11mS or ( 1/ODR + 1mS ).
The calculated time may be better depending on what ODR frequency you set but I found
( 1/ODR + 2mS ) is safer
I should note the doc I got from Terry at FirstAlliance(ST support rep) shows a different flow chart than the public PDF
I asked Terry and you can contact him for the doc but i asked and was told not to post it in the public sector
We’re moving the ST Community to a new platform to give you a better and more reliable community experience.